Giesen WxA (W6) ET question #1252
Answered
by
MAKOMO
redadder22
asked this question in
Q&A
-
Hi, For some reason the ET is displaying the gas % level instead of the ET value. Any idea why this is happening? |
Beta Was this translation helpful? Give feedback.
Answered by
MAKOMO
Aug 18, 2023
Replies: 1 comment 4 replies
-
Most likely that you are reading either the wrong device or port in:
If you want others to be able to help, either post some screenshots of those windows (from the GitHub web page not by email) or share your artisan settings. You can export your settings via Nico |
Beta Was this translation helpful? Give feedback.
4 replies
Sign up for free
to join this conversation on GitHub.
Already have an account?
Sign in to comment
That symbolic assignment Y3 for ET instructs Artisan to take the values for this channel from the first channel of the first extra device. Just remove that Y3 assignment here.